        Пленарные лекции » ...
        PRECISION AND ACCURACY
        IN XRF, TXRF AND SRXRF
        P.S. Hoffmann
        Darmstadt Technical University, Materials and Earth
        Sciences, Chemical Analytics,
        Petersenstr. 23, D-64287 Darmstadt, Germany dg7j hrzpub.tu-darmstadt.de
            Each analytical result has to be
        characterized by an uncertainty. This uncertainty (or error of the
        result) has to be distinguished into “random errors“ and into
        “systematic errors“ (1). To determine a random error usually the
        arithmetic or geometrical mean of the results of repeated measurements
        is used for calculation of the “relative standard deviation (RSD)“.
        This RSD (or the multiple) is a measure for the width of the scatter of
        all data produced at identical conditions and therefore for the
        precision (repeatability) of the result. The determination of the
        systematic error and of the accuracy (difference to the true value) will
        be performed by very different methods:
          Use of two, or better more different, independant
          analytical methods,
          Use of Certified Reference Materials as test
          samples,
          Spiking of samples with analytes,
          Participation in round robins,
          Use of control charts and duplicate plots, use of
          Identity Line plots.
            For the X-ray methods in
        discussion all these methods can be used. But in practice not all can be
        recommended. The reason for the selection of the method of quality
        assurance (QA) is in this case mainly dominated by the advantage of
        X-ray fluorescence analysis (XRF) of non-destructive characterization of
        solid material. For comparison of results for different samples
        absorption and scattering effects (matrix effects) and enhancement
        effects (secondary excitation) have to be considered. The method of
        quantification and of QA to be used in XRF is in practice a function
        only of the sample thickness: infinitely thick specimen,
        intermediate-thick samples, and practically infinitely thin samples (2).
            In XRF-methods the laboratory
        standards, the certified standards and the samples for examination need
        not only a similar chemical composition but additionally similar
        physical properties (e.g., particle size distribution). For infinitely
        thick samples the total matrix effect can be compensated by application
        of the Compton scatter intensity, by internal standardization, standard
        addition, dilution methods, and can be calculated by mathematical
        methods (fundamental parameter, influence coefficient algorithms,
        algorithms with constant coefficients, algorithms with variable
        coefficients, specimen with more than two compounds, algorithms with
        empirical coefficients). For intermediate-thick samples the Compton and
        the Rayleigh scattering intensities ratio is a measure for the total
        mass absorption coefficient. Additionally, the particle effect has to be
        considered in intermediate-thick and thin specimen. Infinitely thin
        samples can be analysed quantitatively by internal standardization, as
        matrix effects and secondary excitation do not play any role.
        References:
        (1) Wenclawiak B.W., Koch M., Hadjicostas E. (Eds.)
        Quality Assurance in Analytical Chemistry, Springer-Verlag, Berlin, 2004
        (2) Van Grieken R.E., Markowicz A.A. (Eds.) Handbook
        of X-Ray Spectrometry, Marcel Dekker, New York – Basel, 2002